Turns on Before Surgery, Won't Turn on After
Posted: Mon May 01, 2023 11:59 am
Some facts:
(A) At ION/UCL we have two transmitters that turned on before surgery, but would not turn on after surgery.
(B) ION reports that several transmitters were running when they checked a few weeks after delivery. Last month, we had two transmitters out of twelve running upon arrival in Paris, after two days journey from Boston.
(C) We check that all transmitters are turned off before and after packing them up in a box.
Our best guess as to what is happening is as follows.
(1) During shipping, the transmitters are subject to a magnetic field so strong that it turns some of them on.
(2) Upon arrival, if we do not check them immediately and turn off any that are running, those that are turned on run down their batteries and turn off. In the weeks that follow, their batteries recover a little.
(3) When it comes to surgery day, the transmitters will turn on briefly, but later they will not run after surgery because what little charge the battery had left is gone.
To stop failures like this we recommend three steps.
(A) We will pack the transmitters in a box that is more like a cube, so as to increase the minimum distance between the transmitters and the box wall. Every extra centimeter between the transmitter and the source of the magnetic field will decrease the probability that the field can activate the transmitter's magnetic switch.
(B) All recipients check their transmitters upon arrival and turn off any that are running.
(C) Before surgery, instead of merely detecting reception, check the average value of the signals, from which we can deduce the battery voltage using: V_bat = 1.8 * 65536 / V_ave. If the battery voltage is less than 2.6V, the transmitter has less than 10% of its battery left.
In the long run, we would like to provide a simple "detector box" you can keep in your office, and in which you can store transmitters. If one of them is running, a light will turn on. A box that is inexpensive enough that you won't hesitate to buy one. We will also consider using a less sensitive magnetic switch.
(A) At ION/UCL we have two transmitters that turned on before surgery, but would not turn on after surgery.
(B) ION reports that several transmitters were running when they checked a few weeks after delivery. Last month, we had two transmitters out of twelve running upon arrival in Paris, after two days journey from Boston.
(C) We check that all transmitters are turned off before and after packing them up in a box.
Our best guess as to what is happening is as follows.
(1) During shipping, the transmitters are subject to a magnetic field so strong that it turns some of them on.
(2) Upon arrival, if we do not check them immediately and turn off any that are running, those that are turned on run down their batteries and turn off. In the weeks that follow, their batteries recover a little.
(3) When it comes to surgery day, the transmitters will turn on briefly, but later they will not run after surgery because what little charge the battery had left is gone.
To stop failures like this we recommend three steps.
(A) We will pack the transmitters in a box that is more like a cube, so as to increase the minimum distance between the transmitters and the box wall. Every extra centimeter between the transmitter and the source of the magnetic field will decrease the probability that the field can activate the transmitter's magnetic switch.
(B) All recipients check their transmitters upon arrival and turn off any that are running.
(C) Before surgery, instead of merely detecting reception, check the average value of the signals, from which we can deduce the battery voltage using: V_bat = 1.8 * 65536 / V_ave. If the battery voltage is less than 2.6V, the transmitter has less than 10% of its battery left.
In the long run, we would like to provide a simple "detector box" you can keep in your office, and in which you can store transmitters. If one of them is running, a light will turn on. A box that is inexpensive enough that you won't hesitate to buy one. We will also consider using a less sensitive magnetic switch.